  • A). Parts of the analytical balance
    1. Examine the analytical balance with its parts and proper operation.
  • B). Performance evaluation of the analytical balance What step is this?
    Warm the analytical balance for 15 minutes while
    placing the standard mass close to the balance in
    order to reach the room temperature while
    stabilizing the balance.
    First Step
  • B). Performance evaluation of the analytical balance What step is this?
    Weigh the standard mass with a tweezer
    2nd Step
  • B). Performance evaluation of the analytical balance What step is this?
    Perform accuracy and repeatability assays by
    making five replicate readings. Remove the standard mass from the pan and tare the balance in between measurements.
    3rd Step
  • B). Performance evaluation of the analytical balance
    What step is this?
    Perform drift assay by placing the standard mass on
    the balance. Obtain the initial reading then leave the
    standard mass for 5 minutes before obtaining
    another reading.
    4th Step
  • B). Performance evaluation of the analytical balance
    What step is this?
    Repeat the process three times (three trials)
    5th Step