Beer’s Law - the concentration of a substance is directly proportional to the amount of light absorbed or inversely proportional to the logarithm of the transmitted light
Beer's Law: Or unknown concentration is directly proportional to the intensity of the color of the solution (not applicable to things we read at 340 nm, which tend to be colorless)
Instrument must be "blanked" or calibrated; this tells the instrument how much light is equal to 0% absorbance or 100% transmittance
When the instrument is being "blanked," reagent itself will absorb some light or water and cuvette will absorb some light. This must be accounted for by setting absorbance of blank to 0%
When the actual patient sample is measured, the absorbance should be less, as the reagent absorbance is subtracted
the incident light in spectrophotometry is the beam of light that is hitting the sample
in the sample - interferences include hemolysis and icterus
depending on the wavelength of light, hemolysis may interfere with all wavelengths
hemolysis will appear reddish
icterus - will affect range of wavelengths, but not entire spectrum
icterus appear greenish-brown
%T is the percentage of light that makes it through the sample
%T is what's actually measured by the photodetector
the graph of percent transmittance vs concentration of the unknown is a logarithmic curve
logarithmic - 3 is 10x stronger than 2
%T is measured directly by spectrophotometer
Absorbance's graph produces a linear curve when absorbance is plotted against concentration
absorbance must be calculated by spectrophotometer, as photodetector only detects transmitted light
The greater the absorbance, the greater the concentration
the greater the percent transmittance, the lower the concentration
I= transmitted light or light detected with sample
Io= incident light or amount of light detected with blank
for every analyte we are measuring, it has its own molar absoptibity
molar absorptivity is defined as the fraction of a specific wavelength of light absorbed by a given molecule
A=ExBxc
E and b are constants, therefore A is proportional to the concentration
wavelength accuracy is checked by using blanks of known maximum absorbance
stray light - wavelengths of light outside the band transmitted by monochromator
we select a wavelength and expect all wavelengths of light to be within band pass
if outside band pass and light still transmitted to photomultiplier when using blanks of known maximum absorbance, this may cause stray light
stray light can be caused by
scratches on the optics
dust particles in the light path
prism may be scratched and cause irregular transmittance
diffraction grating may have imperfection
to fix stray light
blow air in spectrophotometer
check for stray light using cut off filters rather than blank of maximum absorbance
cut off filter - filter blocks all light within bandpass
linearity - change in concentration results in a straight-line calibration curve
how to check linearity on analyzer
using series of diluted colored solutions
linearity on analyzer
series of diluted colored solution
disadvantages: pipettes must be extremely precisely calibrated; very little room for error and technique must be good
linearity of analyzer
most facilities purchase standards that are already prepared
can only check absorbances of the spectrophotometer
most linearity is applied to analytes - use commercially prepared stock solution, create dilutions, and measure concentrations, and interpolate unknown concentration
e.g. glucose
interferences of spectrophotometric reading include hemolysis and icterus and lipemia
turbidimetry is a technique and not an instrument
turbidimetry uses a spectrophotometer to determine the amount of light that is blocked
turbidimetry refers to the light that is blocked in a sample; even if it appears clear or colorless to the naked eye, spectrophotometer reads light that is blocked
e.g glucose hexokinase measures absorbance of light by NADH