Jackson Cross Cylinder

Cards (57)

  • What is the primary purpose of the Jackson X-cyl technique?
    Astigmatism determination without fog
  • What does the Jackson X-cyl technique rely on during the procedure?
    COLC being placed on the retina
  • What is the abbreviation for the cross cylinder used in the Jackson X-cyl technique?
    1. cyl
  • What does the Jackson X-cyl technique utilize for astigmatism determination?
    Comparison of separate images and bracketing techniques
  • What is the design of the X-cyl lens?
    • Positive cylinder on one surface
    • Negative cylinder of numerically equal power on the other surface
    • Axes of the cylinders at right angles
  • What is the spherical equivalent of the X-cyl lens?
    Plano -0.25 +0.25
  • What are the common powers of the X-cyl lens?
    0.25, 0.37, 0.50, 0.75
  • When is the 0.25 X-cyl most commonly used?
    If visual acuity is better than 6/12 with BVS
  • How is the X-cyl lens mounted when used with a trial frame?
    Mounted in a handle at 45° to the axes
  • How is the X-cyl lens marked?
    With the position of the axes, preferably with plus or minus signs
  • What color is usually used to denote the negative sign on the X-cyl lens?
    Red
  • What color is usually used to denote the positive sign on the X-cyl lens?
    White
  • What should be checked regarding the markings on the X-cyl lens in the U.K.?
    They may be opposite, so it's best to check markings
  • What are the components of the phoroptor X-cyl setup?
    • Trial lens cylinder
    • Cross cylinder
    • Adjustment for trial lens cylinder
  • How is the X-cyl used during the examination?
    The lens is 'twirled' or 'flipped' for patient comparison
  • What will happen if the patient is emmetropic during the X-cyl examination?
    Both views will appear the same
  • What occurs if the patient is ametropic during the X-cyl examination?
    One view will be preferred
  • What does axis determination rely on in the Jackson X-cyl technique?
    The effect of obliquely crossed cylinders
  • What does power determination rely on in the Jackson X-cyl technique?
    Changes in the interval of Sturm
  • What is the significance of the mean power of the lens being zero in the X-cyl technique?
    The position of COLC is not affected by the cross cylinder
  • What does the patient view during the axis location determination?
    A circular target equivalent to 6/12
  • How is the X-cyl introduced for axis location determination?
    With axes along 90° and 180°
  • What happens when the lens is flipped or twirled during axis location determination?
    The axes are interchanged
  • What does the patient need to indicate during the axis location determination?
    Whether the circle appeared sharper with the lens in the first or second position
  • What should be noted regarding the preferred position of the X-cyl minus axis?
    It indicates the orientation of the patient's astigmatism
  • What does it mean if the preferred position of the X-cyl minus axis is at 180°?
    The minus axis of the astigmatism lies nearer horizontal than vertical
  • How is the X-cyl introduced for further axis location determination?
    With axes along 45° and 135°
  • What should be noted when the lens is twirled during this determination?
    Which view is better
  • What does the preferred position of the X-cyl minus axis indicate?
    It indicates the orientation of the patient's astigmatism
  • What should be done after determining the patient's minus cylinder axis?
    Introduce a working cylinder at the estimated axis
  • How is the power of the trial lens cylinder determined?
    It depends on visual acuity with BVS
  • What should be used if the visual acuity is better than 6/9?
    • 0.50 DC trial lens cylinder
  • What should be done if the visual acuity is worse than 6/9?
    Work out from retinoscopy, tables, or rule of thumb
  • What must be added during the Jackson X-cyl procedure for every -0.50 DC added?
    A +0.25 DS lens
  • How should the X-cyl be introduced during axis refinement?
    Handle aligned with the estimated axis
  • What happens when the X-cyl is flipped during axis refinement?
    The patient is asked which view is better
  • What does the combination of the trial lens cylinder and the oblique placed X-cyl create?
    A resultant cylinder
  • What should be noted regarding the preferred position of the X-cyl minus axis during axis refinement?
    It indicates the direction for rotating the trial lens cylinder
  • What should be done if the trial lens cylinder needs to be rotated in the opposite direction?
    Reduce the amount of rotation and bracket
  • What indicates that the trial lens cylinder is at the precise axis of required correction?
    Both views will be the same